> I need to set my thinking straight on how to connect to a type-K
> thermocouple...
[quoted text clipped - 7 lines]
> Am I correct in assuming that as long as these are maintained at the same
> temperature they won't skew the actual temperature measurement?
Yes.
Nick. - 28 Dec 2005 16:47 GMT
> Yes.
Blimey, that was quick (and to the point!)
Thanks.
Dan Hollands - 28 Dec 2005 18:08 GMT
>> I need to set my thinking straight on how to connect to a type-K
>> thermocouple...
[quoted text clipped - 11 lines]
>
> Yes.
And the cold junction of the OpAmp must be at that same temperature
Often the plugs and sockets are made out of thermocouple material which
takes them out of the equation. Everything from the point were the
thermocouple material connects to non thermocouple material must be at the
same temperature. Careful arrangement of components to avoid heating or
moving air is required to make a stable accurate device.
Dan

Signature
Dan Hollands
1120 S Creek Dr
Webster NY 14580
585-872-2606
dhollan3@rochester.rr.com
www.QuickScoreRace.com
jibberjabber - 31 Dec 2005 03:18 GMT
A connector/junction/etc only needs to be isothermal across itself. It does
not need to be the same temperature as all the other connectors (if that is
what you were asking). Ex.: a chromel wire enters a copper connector then
exits as a chromel wire. If the 'entrance' junction is the same temperature
as the 'exit' junction then the errors cancel.
If you have a series of junctions then the first junction (where the TC wire
first connects to another metal) has to be the same temperature as the
AD595. If not then you get an error equal to this difference.
Page 5 of AN-274 (Analog Application Sheet) shows inserting a second TC with
"reverse polarity" in series with the first TC to cancel the error. See the
diagram for placement details.
jj